Calibration Not Calculaing Expected Concentration Properly
Posted: Tue Feb 27, 2018 9:31 pm
by sohaib1
I prepared 5, 10, 50, 100 PPM standard from pure ethylene glycol ran it on GC-FID. When i went and did quantification of calibration. The concentrations are not right. Something weird is going on in the software. Can anybody help please. For 5ppm it is giving 17ug/L. For 10PPM it is giving 9.5 ug/L. For 50 PPM it is giving 17ug/L for 100 PPM it is giving 115ug/L. I tried everything i don't know what am i doing wrong. Can anybody help me in this situation. Thanks
Re: Calibration Not Calculaing Expected Concentration Properly
Posted: Tue Feb 27, 2018 10:03 pm
by EmpowersBane
Could be a lot of things. Did you populate your component editor correctly? Did you maybe use the dilution/sampleweight fields when you shouldn't have? Typically the concentrations are input into the component editor.
Also, did you set up a sample set with wizard, are you sure you set the standards to different levels per vial instead of the same level?
What have you tried?
Posted: Thu Mar 01, 2018 1:31 pm
by ydna1977
Agreed
Check calibration standards and curve linear
Go to alter sample table and check component table
Create new processing method if nothing obvious
Posted: Tue Mar 06, 2018 2:49 pm
by sohaib1
Thank you very much for your responses got it working now it seems my calculations were wrong
Posted: Wed Mar 07, 2018 12:20 pm
by ydna1977
what was the issue here ?
It might be useful to share your miscalculation in empower