We have a Waters Quattro premier + UPLC.
Mobile phase is 75% H2O, 25% MeCN +0.1% Formic acid. 0.3ml/min
Column is an Aquity C18
Runtime is 16minutes - I know this is long, but we are separating isomers.
The problem is that over a series of injections of compound diluted in mobile phase, we get a gradual loss in sensitivity. If 100 samples are run, the last one will be roughly 10% of the first.
When the machine is left to run mobile phase over night without injecting, the sensitivity returns on re-injection next morning, so it is not a stability problem.
Reading data from the UV absorbance detector has good consistancy, showing it cant be the injection system.
We added an internal standard (IS sensitivity also drops), however the massive loss in sensitivity means our method would not be suitable for analysis of biological extracts, we need to keep the sensitivity up!.
We have also tried diverting the first 5minutes of the run to waste, incase there were any impurities fouling the MS, but that hasnt helped either.
Any ideas would be appreciated.