Vitamin D3 determination in lanoline ointments.
Posted: Tue Feb 27, 2018 3:02 pm
I have been working on the determination of cholecalciferol (Vitamin D3) in an ointment (Lipidic ointment based on lanoline). For the extraction of the vitamin, the following saponification process has been tested:
1. Weigh 4 grams of the ointment in a erlenmeyer flask with screwed cap.
2. Add 20 mL of ethanol and 16 mL of KOH 5N
3. Proceed to cover the flask with the cap, and put into a thermoregulated bath at 50°C and 70 rpm for 60 minutes.
4. When the saponification process have been concluded, add 20 mL of water and 20 mL of n-Hexane (HPLC Quality).
5. Wait until decantation and take a portion of the top layer of the 2 phases that just formed.
6. Filter the solution through a 0,22 filter and inject in the hplc
Column: Kromasil Si- 250 mm x 4,6 mm x 5um
Column temperature: 40°C
Injection Volume: 50 uL
Flow: 1,0 mL/min
Mobil phase: Hexane/Isopropanol [99:1]
The problem that we have here, is that we cannot obtain a clean chromatography, and the peak of vitamin D3 is impure, the reason of that is the matrix effect of lanoline (which presents many peaks). So, I want to know if any of you have been working with something similar.
Flow: 1,0 mL/min