but once again how can you explain that I have no problems on varian system;
moreover, I tried yesterday other fittings (peek tubings with 'universal' fittings), which I usually use on varian system, the problem did not disappear
As previously said by Chromatographer1, having not enough dead (or system dwell) volume in Agilent 1200 system cannot provide appropriate mixing of the sample injected and the mobile phase (but peak splitting will result only if your sample solvent is stronger than the initial composition of the mobile phase)
I can recall that Agilent 1200 has system dwell volume of about 0.5mL, but if you have the Rapid Resolution configuration the system dwell volume may be less than that (< 0.5mL)
I don't know what is the system dwell volume of your Varian HPLC system.
furthermore, when I was working couple months ago in different company also using agilent systems, there was no such problems, I was using there columns of different manufacturers and peek tubings with appropriate fittings;
Yes, but tell us did you use the same method/same assay then as now?
Did you use Agilent 1200 Series HPLC system in your previous company or other older Agilent models?
i have tried regural tricks (changing injection volume, different concentrations etc) but it did not help
Ok, then probably you're not facing with this "dilution" problem of your sample, but still can you provide more information about the solvents used for preparation of your sample and your mobile phase composition?