- Posts: 3
- Joined: Mon Jan 22, 2018 2:17 pm
(Sudan dyes need ms/ms detector) diffecult in DAD OR PDA
and am little confused


Agilent 6470 or 6460
Water xevo TQD (or s micro )
shimadzu 8050
sciex 5500 or 4500
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
thx for your advice and u are 100% right but all application is not difficult to analyzed with all those brands and there are a lot of applicant note for them in those brands so i ask about software ,hardware ,stability and troubleshootingNot "UPLC". Wrong term as that is a trademark belonging to Waters Corp. You mean UHPLC or HPLC (UHPLC is HPLC). If you call it HPLC, you will always be using the correct terminology. UHPLC was just created as a marketing name. Here is a link on the web which will explain it to you (https://hplctips.blogspot.com/2015/08/t ... lc-or.html )
Most of the big brands are good, but you should put together some test samples (3-5 know samples with methods that you are familiar with, not unknowns) and arrange to have demos of the different systems at the vendors application lab location to try them out. Be sure and study the software (In some ways the selection of software is as or even more important than the hardware) and certainly bring 1 or 2 other people with you who have many years of professional experience using and troubleshooting HPLC systems (any brands).
Asking people on a group like this for advice on which system you should buy is not the best approach. These people are not you. They do not run your samples. They do not have the same training as you or have the same needs as you. Groups can offer advice regarding their personal experience with service contracts or long term problems, but not on what is best for you.
thx a lot u really help meHello Alye,
I have used the instruments of two of these three vendors you list, Waters, and Agilent, although those LC/MS systems I've used were both single quads, not QQQs.
I found the training sessions of both of these vendors to be excellent, probably I'd give Agilent a slight edge.
Where I am in the U.S., both of these vendor's support is excellent--can't say what it is like where you are located. Training and troubleshooting are also very good for both Waters and Agilent.
In my opinion, it is easier to perform your own service on Agilent's LCs, at least at first until I gained experience with Waters Acquity systems. For the ESI work, Agilent's OpenLab was easier for me to learn than MassLynx as OpenLab was similar to ChemStation.
In my opinion, both the Waters and the Agilent LCs worked well enough (Acquity-I and 1290)...with 2.1 mm ID columns...and both LCs have their idiosyncrasies.
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.