we observed the following phenomena:
We adopted a method from a project partner, and there the reference was dissolved in 100%MeOH, while the inital conditions of the gradient were 15%ACN/acidified water; column is a Nucleodur C18, 5µm, 4.6x250mm; InjVol was 20µL.
We first run the method on our Waters Alliance 2695 and on the 600-system with 717 injector (both low pressure gradient), and got "normal" chromatograms with good peaks.
Now we did a transfer to another lab which is using older Varian (low pressure) and new Agilent 1200 RRLC (high pressure gradient) systems and observed very bad peak distortion when injecting >10 µl of the reference solution on both systems.
Diluting the reference solution to 50%MeOH solved the problem but I wonder why we don't get this phenomena on the Waters HPLCs? (maybe they are simply too good
Does anybody have an idea why the different systems are behaving like this?