Here's my info:
Solvent A: 0.1% acetic acid buffer titrated to pH 3.5 with NaOH
Solvent B: acetonitrile with 0.1% acetic acid
Flow rate: 1mL/min
Column: Agilent Zorbax SB C18 4.6x250mm 5 micrometer particle size
The method is:
0-2 min: 5% B
2-14 min: 15% B
14-19 min: 18% B
19-32 min: 35% B
32-33 min: 60% B
33-35 min: 5% B
I've ran this a few times and all I see are big peaks right around 5 minutes at the beginning of the run, and they fade away into smaller peaks. I was wondering if this was from my standard but then I ran it with a vial of water instead of my standard and I saw the same peaks. The standard I injected was 0.1mg/mL cyanocobalamin, and I injected 5 microliters. When I ran the plain water I also injected 5 micrometers. Pictures are below.
Cyanocobalamin try 1:

Cyanocobalamin try 2:

Water (Ran directly after cyanocobalamin try 2):

I'm also noticing pressure fluctuations during the run of about 5-10 bar. The pressure decreases generally as the percentage of pump B increases. At the beginning it might go from 74 to 79 bar over the course of 6 seconds, and at 60% B I will be going from 39-50 bar over he course of a few seconds.
Not sure if it means anything, but when I'm priming the pump after turning it on, and I'm using 100% A, I can see the solvent/ little air bubble (It forms after turning the machine off for the day in the clear tubes going into both outlet valves.) going into the active inlet valve on pump B actually getting pushed backwards a small amount after each pump. After switching to 100% B to prime that side this air bubble eventually goes away and comes out the purge valve and the pump runs normally.
There's no leaks that I can see, outlet ball valves were recently changed, piston seals looked good, and were changed on one side where needed.
It's very strange because this machine was working before I changed the column, and solvents. It successfully separated and detected formate but this method only used one solvent. Any help or insight is greatly appreciated!