Calculating LOD/LOQ (regression line)
Posted: Fri Apr 11, 2008 6:44 pm
I'm banging my head against a wall here trying to figure this out. I'm trying to calculate LOD/LOQ for an HPLC method using the "The residual standard deviation of the regression line" method. I have the following data:
Nominal Area Conc.
0.50 0.01285 0.515
1.00 0.02065 0.969
2.00 0.03845 2.005
5.00 0.09020 5.019
10.00 0.17560 9.992
Slope: 0.017173139
Intercept: 0.004009386
RSQ: 0.999973532
Whichever way I try and calculate LOD/LOQ, I end up with a number mid-curve. With this method the LOD/LOQ should be WAY below 0.5.
Can anybody do a step by step Excel formula on how to calculate LOD/LOQ with this data using the "residual standard deviation of the regression line" method.
Thanks in advance.....
Nominal Area Conc.
0.50 0.01285 0.515
1.00 0.02065 0.969
2.00 0.03845 2.005
5.00 0.09020 5.019
10.00 0.17560 9.992
Slope: 0.017173139
RSQ: 0.999973532