To be honest, the FID ADC measures voltage got from output of low-current amplifier.
None of the commercial instruments were calibrated with real low-current source (like this
http://ieeexplore.ieee.org/document/4277008/ or like this
http://www.transmille.com/ea013-picoamp-source-adpater) to be checked these pico-amperes are real.
So it is a matter of trust of you to the vendor who provide you with pico-amps or mV which is proportional (according to the Ohm law I=U/R) to the current which is proportional to charge actually flown in ionization chamber which is proportional to ionization effectivity of your H2 flame and slightly relate with substance structure.
Also none of the vendors provide you with the feedback resistance (R) in low-current amplifier because in some cases (like Agilent) it is pure virtual thing.
Better take known amount of substance, inject it and get straight or relative response of FID (measured in whatever units).