Acetic Acid Method Development
Posted: Wed Apr 09, 2008 11:04 pm
Hi Guys,
I am having to develop a GC method for residual solvents for a particular API. Of the six solvents, one is acetic acid. The client is adamant about having one method.
I am steering clear of headspace and am using liquid injection. The material is not water soluble to a great extent, so I am using DMF as a diluent.
This is a 1µL injection onto a 624 phase 30m column on an Agilent 6890. I am having to play with solvent concentration levels so as to be able to detect 1,2-dimethoxyethane (one of the other solvents) and have a reasonable test sample concentration. In adjusting the concentration levels, I am having to adjust my split ratio to maintain a reasonable response for the DME.
Now the interesting part: while doing all this, I have noticed acetic acid elute later as I lower the split ratio. The ratio went from 15:1 to 2:1 and I have seen almost a 2.5 minute shift in retention time. None of the other five solvents moved.
Now, not having much experience with acetic acid by GC, but knowing the 624 column is not the best for acetic acid, has anyone had this issue before? Let's just say for arguments sake that the 624 column is my only choice, what is the explanation for the retention time shifts?
I have tried a wax column and a G27 column and both gave me partial co-elution of 1,2-DME and IPA. This is no good. The 624 gave the best separation.
Burt
For those who are wondering, the solvents are: methanol, ethanol, IPA, acetone, 1,2-DME, acetic acid. API is marginally soluble in water, soluble in acid (but is unstable), soluble in DMF, DMSO.
Burt