Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
We are running EPA method 508 for pesticides on 5890's and are having a recurring problem. We do all the standards, degradation check, etc, but when we get to the CCV - the results are too high. Chromatography and RT's are right on and the ICV passes fine - just the CCV. I am not sure if it is an inlet problem or something on the detector end. We do pesticides standards, chlordane, and pcbs at the same time. I kind of remember 10 years ago or so, that I would do pests and then do pcbs to avoid a similar problem - but I honestly cannot remember. Could there be a problem with make up gas flow rate? Any input would be appreciated. Thanks :)

Barneyjo,

Are you using an internal standard? My experience has been that the 5890 ECD had a signficant amount of drift over time as the oven top plate becomes hotter and hotter. I think this has to do with the overall detector temperature getting higher over time. To test this, put a long delay in between preliminary runs like breakdown and your CCV in order to let the top plate cool off. Also make sure that your detectors are well insulated. If you can run an IS this will compensate somewhat for this phenomenon.

I will be curious to hear more about your plight.

Best regards.
2 posts Page 1 of 1

Who is online

In total there are 102 users online :: 0 registered, 0 hidden and 102 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 102 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry