Hi everyone,

We are using HPLC 8453 UV- visible Spectrophotometer which is a single-beam to monitor the pH of seawater over days. We use three wavelengths 730nm, 578nm and 434nm to measure and record the absorbances, in which 730nm is the reference wavelength. But the absorbance of reference wavelength is not consistent over time (hours or days), usually we got the increasing trend or fluctuation curve (under the condition of no CO2 exchange with the sample and no biological activity within the sample). The lamp of spectrophotometer is Ok because we got all tests passed (self test, lamp intensity, lamp stablity) executed by operating the software.
Could anyone explain to me the cause of the inconsistency of non- absorbing wavelength (730nm) over time ?

Many thanks in advance,

I would speculate that you are seeing fluctuations in lamp output (single-beam spectrophotometers do not correct for this).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 10 users online :: 1 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry