HP 5890 step down transformer requirements
Posted: Fri Feb 09, 2018 5:33 am
by petehops
Hi all
I am in Australia and have had a HP5890 GC come into my possession from a colleague in the US which is 120V. I need to get a step down transformer to be able to operate it, any ideas on the output in watts i would require?
Many thanks in advance
Re: HP 5890 step down transformer requirements
Posted: Fri Feb 09, 2018 8:38 am
by Peter Apps
There should be a panel on the back with the power requirements. Also, if I remember correctly (which is hope over expectation !), you may be able to convert it to 220 V by shifting some jumpers. Do you have manuals ?
Peter
Re: HP 5890 step down transformer requirements
Posted: Fri Feb 09, 2018 9:52 am
by dblux_
Hi all
I am in Australia and have had a HP5890 GC come into my possession from a colleague in the US which is 120V. I need to get a step down transformer to be able to operate it, any ideas on the output in watts i would require?
Many thanks in advance
This should be very useful:
https://www.agilent.com/cs/library/supp ... a15282.pdf
Re: HP 5890 step down transformer requirements
Posted: Fri Feb 09, 2018 8:44 pm
by James_Ball
I know the 5890s required 110v and 20amp circuits if that helps. I believe they pull 2200VA max as listed on the back panel, but has been a while since I saw one.