GPC Calibration
Discussions about gel permeation chromatography / gel filtration chromatography / size exclusion chromatography
8 posts
Page 1 of 1
What are some ways to do internal calibration on Gel Permeation Chromotagrapy? We run narrow molecular weight standards every once in a while to generate a calibration curve but is there a way to do instrumental calibration? Just to make sure the instrument itself is running properly? Would reconditioning the column be a part of it? Any help will be grealty appreciated. Thanks
-
- Posts: 267
- Joined: Tue Oct 12, 2004 9:35 pm
Other than making sure your column is clean, which will be covered by your column manual, running standards is about all there is.
As for the instrument, do the proper preventive maintenance and performance checks. Check the pump flow rate periodically. Check that the injector is precise. Check that the detectors are performing. Also, just make sure that the tubing volume on your instrument is appropriate. This usually isn't a concern with GPC columns that are 300 x 7.8 mm with 7-10 micron particles.
As for the instrument, do the proper preventive maintenance and performance checks. Check the pump flow rate periodically. Check that the injector is precise. Check that the detectors are performing. Also, just make sure that the tubing volume on your instrument is appropriate. This usually isn't a concern with GPC columns that are 300 x 7.8 mm with 7-10 micron particles.
-
- Posts: 14
- Joined: Wed Nov 28, 2007 8:04 pm
Thank You Noser222!
Would performing a plate count to check for column efficiency fall underneath this? Is that something typically done to ensure column efficiency on a relative regular basis or just cleaning the column would suffice? I am currently using Waters Ultrahydrogel columns.
Would performing a plate count to check for column efficiency fall underneath this? Is that something typically done to ensure column efficiency on a relative regular basis or just cleaning the column would suffice? I am currently using Waters Ultrahydrogel columns.
-
- Posts: 267
- Joined: Tue Oct 12, 2004 9:35 pm
Yes, tracking plate count of a standard and pressure are a good idea.
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
Since in GPC all data are based on accurate retention times, which means accurate flow, you can check the system by injecting regularly the plate count marker - not so much for the plate count, but for the retention time.
-
- Posts: 14
- Joined: Wed Nov 28, 2007 8:04 pm
Thanks guys.. really appreciated it
-
- Posts: 405
- Joined: Wed Dec 05, 2007 4:55 pm
One final comment: as noted, retention time is the primary separation variable in GPC. You could simply re-analyze your standard as a sample, and verify that the retention times of the components have not changed. You could put fairly tight specs on the retention time windows, and that would give you more confidence in your results.
But also you should realize that GPC calibrations are log-based, and a 0.1 min change in retention time can mean a MW change of about 20,000 units in some cases. If that difference is important, then you need to be more careful with your system.
Finally, the best way of minimizing this problem is to add an internal standard to your sample. For example, dissolve your sample in a solution of solvent that contains a small, retention time marker (small molecule that elutes at the end of the chromatogram, after your peak(s) of interest). Most data systems have an internal standard feature that will calibrate and calculate the "ratio" of retention times (peak/internal standard), which corrects for minor flow rate changes.
The only concern is that this internal standard is sufficiently resolved from any other baseline changes (common in THF systems for example) that occur in that time region. But you only need a defined peak maximum, not a fully resolved peak.
But also you should realize that GPC calibrations are log-based, and a 0.1 min change in retention time can mean a MW change of about 20,000 units in some cases. If that difference is important, then you need to be more careful with your system.
Finally, the best way of minimizing this problem is to add an internal standard to your sample. For example, dissolve your sample in a solution of solvent that contains a small, retention time marker (small molecule that elutes at the end of the chromatogram, after your peak(s) of interest). Most data systems have an internal standard feature that will calibrate and calculate the "ratio" of retention times (peak/internal standard), which corrects for minor flow rate changes.
The only concern is that this internal standard is sufficiently resolved from any other baseline changes (common in THF systems for example) that occur in that time region. But you only need a defined peak maximum, not a fully resolved peak.
Merlin K. L. Bicking, Ph.D.
ACCTA, Inc.
ACCTA, Inc.
-
- Posts: 6
- Joined: Wed May 17, 2006 4:57 pm
Do you work in a regulated testing environment, like a pharmaceutical company? If so, our firm uses an approach that looks at instrument performance and documents it over time in order to assure regulators that all is working well with your system. See link:
http://pssgpcshop.com/easyvalid.htm
http://pssgpcshop.com/easyvalid.htm
8 posts
Page 1 of 1
Who is online
In total there are 5 users online :: 0 registered, 0 hidden and 5 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 5 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 5 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science