alicee! » Sun Apr 06, 2008 3:11 pm
here what you can try on columns you might have:
1. Regular bare silica column
ACN/water/any acid or buffer=90/10/0.1 HILIC mode. Play with pH and ACN to adjust selectivity, retention and peak shape.Watch for sulubilty of buffer, can try ion-exchange on silica column at low ACN concentrations. If retention is too long reduce amount of ACN and add more buffer
2. Regular C18 or C8 column
0% ACN with sodium phosphate pH 7. You will be using residual silanols on the column to retain your amino sugar by ion-exchange mechanism on weak and limited amount of acidic sites of the column. Use buffer concentration and pH to control peak shape and retention. This might be tricky on the long run as slight variation is silanol activity/amount will cause shift in retention time (which might be dramatic)
3. Any C18 column with HFBA or other ion-pairing reagent. this might be tricky because of RI detection
4. Derivatization of amino groups to create chromophors and add hydrophobicity (benzyl chloride). This might be tricky in terms of obtaining quantitative accurate results, because you hav to push this reaction to compliton
1. u mean ion-exchange using the NH3+ group? as oh- will be present on silica right? yes i shal try that..
2. i have tried only water.. that hasn't worked.. so i doubt if phosphate buffer with pH 7 would work.. i can try that too.. what say?
somehow i feel with 80% methanol it was bound.. but then i m not sure? could u explain that? is there a possibility?
3. i m not thinking of trying that for now. shal try it if nothing works.
4. i have tried OPA derivitization.. but i would face the problem in quantitation as you say.
besides, i m planning to try an anion exchange.. my aminoglycoside is kanamycin.. so i may try anion exchange with pH 4-10 gradient.. any experiences with that?
would RI be very less sensitive than ECL detector? any ranges? i need a more sensitive mehtod. in that case ECL / derivitization would be better?
thanks a ton!
importantly, i have understood the importance of pH or ion-pairing agents here.. but ACN.. i fail to understand.. could u help that? thanks!