UPLC conversion

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Dear Friends

I have developed a HPLC method of test concentration 0.01ppm with response 39000. I have tried to adopt the same method to UPLC to get better repsonse. But i have not succeeded in that more over i am getting the 18000 response only. What is reason behind this. I am looking forward for your advise.

Regards
Chand

I hate to nitpick, but there is no way to comment intelligently without a lot more information.

First of all, more detail would be helpful: what size column, flow rate, injection volume did you start with, and what are you using with the UPLC system?

Second what do you mean by "response" (area? peak height? what units?)?

Third, what really matters as far as sensitivity is concerned is signal/noise ratio. How does that compare between the two systems?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 24 users online :: 3 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Bing [Bot], Majestic-12 [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry