we have a typical situation here, we are tring to do linearity of an Active substance from 15µg/mL to 630µg/mL, surprisingly the linearity is not passing , the square of correlation coeffcient is around 0.9966( Limit NLT 0.999), we have repeated the experiment and ending up with almost same result ,iam giving response against Concentration
Conc( µg/mL) Response
15.738 97063
23.607 144072
47.213 281772
94.426 544904
118.033 670788
236.066 1236063
393.443 1908790
472.132 2213992
550.82 2510956
629.509 2811985
we can not change the column as we are verfying this method at our Lab,
Chromatogrphic Conditions :-
Flow :- 1.5ml/min
Temp: 30°C
Column : Symmtry C18 150x 3.9 mm , 5µ
Wavelength : 220 nm
the Sample is Dextroamphetamine sulfate,
Surprisingly linearity is fine when you consider from 15.738 to 236 and it is fine from 236 to 629 , why it is not linear from 15 to 630µg, the peak is not saturated at 220 nm , the height of peak at 220nm for 630µg/ml is only about 0.4AU., please somebody help me in this issue.
thanks