Flow: 0.3 ml/min
Column temperature:65 celcius
Wavelenght: 250 nm
Buffer: 6,25 g SDS and 2,73 g monobasic sodium phosphate dissolved in 500 ml water.Ph is adjusted to 6,8 with %2 NaOH.
Mobile Phaes: Buffer: ACN:MeOH (55:25:25) mixed by volume, and pH is adjusted to 4,2 with phosphoric acid.
Run time: 45 minutes.
You see the usage of high amount of SDS. After 10 injection , one has to clean the column. Thus one analysis is so much time consuming eventhough it is UPLC method.
In my opinion, SDS should be replaced with ion pair agent like Heptane sulfonic acid. Also mobil phase's pH is adjusted to 4,2 after mixing ACN and MeOH. Such a pH adjustment prone to errors since pH adjustment is not measured in aqueous medium.
Am I right about my comment? I want to know your comments on the methods.