Hello again,
If you can please open the following link you can see my chromatogram
http://tinypic.com/r/feklms/9
I do not know if the quantification after 150 times for the left peak is acceptable or not? or should I do quantification based on 100 times dilution?
My diluent is water 100 %
my eluent is:
A: H2O 100 % + 20 mM acetic acid
B: MeOH 100 % + 20 mM acetic acid
In addition, if you open the following link
http://tinypic.com/r/28a3o86/9
You can see my calibration curve based on the concentration I made below:
0.0975 ppb, 0.195 ppb, 0.39 ppb, 0.78 ppb, 1.56 ppb, 3.125 ppb, 6.25 ppb, 12.5 ppb, 25 ppb, 50 ppb, 100 ppb, 200 ppb, and 400 ppb
So, do you think my range is not spread well? One of the main reason I am using many points in lower concentration is that I need to determine the limit of detection and quantification for my method validation. What do you recommend me to do? If my unknown sample specific analyte concentration falls into the middle of this calibration curve do you think it is acceptable?
Thank you for your reply