I am using a GCxGC-ToFMS system. Unfortunately, the operating software does not allow to have a constant flow mode. So, I can only use pressure ramps. I use HP flow calculator to estimate pressures at min and max temp for a given flow rate (1.5 ml/min) for each segment of columns (transfer line, 2nd column, modulating deactivated column, 1st column). Later, I simply divide the difference of press at min and max temp by total run time to figure out the pressure ramp rate.
I am just wandering if I am doing this properly or there may be other approach?