I've noticed that in a method I made, the tailing of the main peak is affected slightly by column temp. 40° T=1.1, 25° it's 0.84. I'm aware that column temp affects selectivity, but why is it also affected tailing? I thought tailing was mostly a function of pH. ty
Your forgetting the other reaction! This is the interaction between the column and your target molecule, that is the chromatographic separation. This reaction is more efficient at a higher temperature (up to a point). Thus you would see a sharper peak.
if TF is < 1, the peak is not "tailing", it's actually "fronting". One possible cause of fronting is limited solubility in the mobile phase; another is a tendency of analyte molecules to form aggregates. Both are less problematic as temperature increases. You can speculate about the specifics, but the bottom line is that temperature can and does impact peak shape.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
3 posts Page 1 of 1

Who is online

In total there are 14 users online :: 0 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry