A number of questions here:
first of all, was this a gradient or isocratic separation? (From context, I'll assume that the separation in question was a gradient.)
second, when you say that only one compound shows the RT variation, are you running several compounds in the same run?
next, (again assuming a gradient) are you equilibrating for
exactly the same time before every run?
(again assuming a gradient) are you using a high-pressure ("two-pump") or low-pressure ("one-pump") mixing system?
finally, approximately what are the retention time and peak width of the peak in question?
Here's what I'm getting at with all of these question:
1. If you have a relatively large column, with long retention and wide peaks, a 6-second variation may, in fact, not be a significant problem.
2. No gradient system generates a perfectly smooth gradient; there are always fluctuations around the average composition. Low-pressure mixing systems (on average) seem to be more prone to this problem, but it's always there so some extent.
3. Some compounds are more sensitive to mobile phase composition than others. That means that small fluctuations in composition may make some peaks move quite a bit while others are almost unaffected. In general, higher molecular weight compounds are more likely to move. The same thing applies to stationary phase equilibration.
All that said, the first thing I would do is to run a "stair-step" test. There is a brief write-up in the
Troubleshooting Wizard on our web site (
http://www.lcresources.com/resources/TSWiz/hs450.htm).
If you're running a narrow-range gradient (e.g., 20 - 40%), you may be able to sidestep the problem by putting your gradient start and end solvents in the A and B reservoirs and running from 1 - 100%.