OK, I understand now: here is the problem: HILIC works with mobile phases of the opposite composition than reversed-phase. You want to work with a LARGE fraction of acetonitrile, and a small fraction of water. You do not need ANY salt for your analytes. I am also not convinced that the HILIC column that you have is the best choice for the task, but we will see after you use the correct mobile phases.
As I said above, use 80% acetonitrile 20% water. Nothing else in it. Use a low wavelength, 210 nm. equilibrate you diol column with this mobile phase until you do not see any drift any more in the baseline. Then inject the three analytes. If they elute too early, go to 90% acetonitrile and 10% water. No salt needed.
If you do not get retention under these circumstances, you need to get a better HILIC column. My recommendation is the Atlantis HILIC silica column from Waters. Start with the same mobile phase as mentioned above (80% acetonitrile).