RT variations
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
9 posts
Page 1 of 1
Has anyone ever had a RT that fluctuates in both directions from one injection to another (up to 0.1 min)? What could cause this ? The autosampler has been checked and the synchronization with the MS acquisition is fine. It's related to one compound and has happened using 2 different columns, no pb with other compounds... Any advice will be VERY appreciated !
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
First: 0.1 min is not that bad at all. Second: The fact that you only see the variation for one compound and not for another tells me that these 2 compounds elute under different conditions (e.g. the first under isocratic and the second under gradient conditions).
If you really want to do something about it, you should focus your efforts on the pump/s, gradient valve (GPV) etc.
Best Regards
If you really want to do something about it, you should focus your efforts on the pump/s, gradient valve (GPV) etc.
Best Regards
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 3
- Joined: Wed Mar 19, 2008 10:42 am
+/- 0.1 min is not acceptable in terms of GLP validation
The other compound was analyzed under isocratic condition as well so I'm 100% sure the pump and valves are working fine
The other compound was analyzed under isocratic condition as well so I'm 100% sure the pump and valves are working fine
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
A number of questions here:
first of all, was this a gradient or isocratic separation? (From context, I'll assume that the separation in question was a gradient.)
second, when you say that only one compound shows the RT variation, are you running several compounds in the same run?
next, (again assuming a gradient) are you equilibrating for exactly the same time before every run?
(again assuming a gradient) are you using a high-pressure ("two-pump") or low-pressure ("one-pump") mixing system?
finally, approximately what are the retention time and peak width of the peak in question?
Here's what I'm getting at with all of these question:
1. If you have a relatively large column, with long retention and wide peaks, a 6-second variation may, in fact, not be a significant problem.
2. No gradient system generates a perfectly smooth gradient; there are always fluctuations around the average composition. Low-pressure mixing systems (on average) seem to be more prone to this problem, but it's always there so some extent.
3. Some compounds are more sensitive to mobile phase composition than others. That means that small fluctuations in composition may make some peaks move quite a bit while others are almost unaffected. In general, higher molecular weight compounds are more likely to move. The same thing applies to stationary phase equilibration.
All that said, the first thing I would do is to run a "stair-step" test. There is a brief write-up in the Troubleshooting Wizard on our web site (http://www.lcresources.com/resources/TSWiz/hs450.htm).
If you're running a narrow-range gradient (e.g., 20 - 40%), you may be able to sidestep the problem by putting your gradient start and end solvents in the A and B reservoirs and running from 1 - 100%.
first of all, was this a gradient or isocratic separation? (From context, I'll assume that the separation in question was a gradient.)
second, when you say that only one compound shows the RT variation, are you running several compounds in the same run?
next, (again assuming a gradient) are you equilibrating for exactly the same time before every run?
(again assuming a gradient) are you using a high-pressure ("two-pump") or low-pressure ("one-pump") mixing system?
finally, approximately what are the retention time and peak width of the peak in question?
Here's what I'm getting at with all of these question:
1. If you have a relatively large column, with long retention and wide peaks, a 6-second variation may, in fact, not be a significant problem.
2. No gradient system generates a perfectly smooth gradient; there are always fluctuations around the average composition. Low-pressure mixing systems (on average) seem to be more prone to this problem, but it's always there so some extent.
3. Some compounds are more sensitive to mobile phase composition than others. That means that small fluctuations in composition may make some peaks move quite a bit while others are almost unaffected. In general, higher molecular weight compounds are more likely to move. The same thing applies to stationary phase equilibration.
All that said, the first thing I would do is to run a "stair-step" test. There is a brief write-up in the Troubleshooting Wizard on our web site (http://www.lcresources.com/resources/TSWiz/hs450.htm).
If you're running a narrow-range gradient (e.g., 20 - 40%), you may be able to sidestep the problem by putting your gradient start and end solvents in the A and B reservoirs and running from 1 - 100%.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 40
- Joined: Wed Aug 01, 2007 11:58 am
Tom
Fred wrote
"The other compound was analyzed under isocratic condition as well "
I conclude from this that it was rather isocratic analysis.
Fred
Are you using buffered mobile phase? Some compounds give fluctuation in retention time in unbuffered conditions.
Fred wrote
"The other compound was analyzed under isocratic condition as well "
I conclude from this that it was rather isocratic analysis.
Fred
Are you using buffered mobile phase? Some compounds give fluctuation in retention time in unbuffered conditions.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Oops!I conclude from this that it was rather isocratic analysis.
That said, if you are set up for "on-line mixing" (i.e., having the pump control the mobile phase composition), the same comments and suggestions would apply.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 1233
- Joined: Tue Oct 11, 2005 7:20 am
I'd agree about the buffered mobile phase, and also suggest ...
- does the variation affect a pure standard of the compound as well?.
- is the sample in a solvent similar to the mobile phase, including any buffer?
- is the sample clean?, as extraneous junk can affect RT.
Please keep having fun,
Bruce Hamilton
- does the variation affect a pure standard of the compound as well?.
- is the sample in a solvent similar to the mobile phase, including any buffer?
- is the sample clean?, as extraneous junk can affect RT.
Please keep having fun,
Bruce Hamilton
-
- Posts: 3
- Joined: Wed Mar 19, 2008 10:42 am
Thanks to all for your replies. I solved the problem, it was the binary pump (Agilent 1100) that was not delivering a constant composition (!). I found that by using only one channel and one bottle of solvent containing A & B. Anyone having met this problem before ?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
It's not that uncommon. The same kind of problem was discussed in one of John Dolan's LC Troubleshooting columns in LC-GC:Anyone having met this problem before ?
LC-GC 24(7) 662-668 (2006)
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
9 posts
Page 1 of 1
Who is online
In total there are 29 users online :: 1 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: rrrrr and 28 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: rrrrr and 28 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science