i have some questions about tubing in the agilent/varian hplc;
i am an analytic chemist at pharmacy company; until now i was warking on varian hplc system; there were peek tubing connections 0.02 in (i.e. 0.5 mm?) just before and after column; the rest of connections were steel and, according to service (technical support), were not 0.02 in but 0.01 in.
since couple of weeks we have agilent 1200, and the connections are steel and 0.17 mm (0.007 in);
because I make analytical methods and validate them, it is very important to me that transfer of the methods fom agilent to varian was not problematical (I mean retention times of the peaks mainly, but shape and symmetry of the peaks is very important as well);
so I called the agilent service and asked what to do, thay suggested changing of all connection tubings to 0.02 in;
I was working couple of months ago with agilent 1100, and I do not really remember we changed all connection tubings; we changed only the tubings before and after column, but i do not remember if the change was so drastic ( I mean from 0.007 in. to 0.02 in.); I think that we change those tubings 0.007 in. to 0.01 in.; and i had never changed the rest of the tubing using agilent 1100;
when I asked varian service, they told me, that it is not really important to change all tubings to the same diameter
so now I am really confused about the situation; first of all what to do with the tubing connections to make method universal for varian and agilent hplc systems
the second thig is the diameters of tubings; is it true that they all need to be the same diameter; is it dangerous for hplc system to use different diameters of the tubings? can I damage the system using other diameter tubings only before and after the column?