JackC » Thu Mar 27, 2008 3:31 am
One thing you need to consider is the selectivity of those columns. For example, the DB-5MS is a silphenylene type of column (where the phenyl is inserted in the backbone of the phase) versus the pendant phenyl types of the Rtx-5MS and HP-5MS. While the vendors do indeed try to "tune" the selectivities to be the same for these columns, they aren't. I can give you two examples: you can separate m- and p-xylene on the silphenylene column (admittedly with a long run time), but you can't do this on the 5% phenyl pendant column. The other example involves separating 2378 dioxin and 2378 furan from interfering congeners with a 5% silphenylene column; can't be done with a 5% pendant phenyl.
Don't forget the Rxi-5ms from Restek, which is new and advertised to be extremely inert, which may be important for your drug analyses at lower levels. Also, consider a guard column, either attached with a connector, or integral to the analytical column, based on your sample description.