Dear Chemists,
I'm experiencing a problem involving Lactic, Malic, fumaric, and Succinic acid using LC/MS ESI with a Phenomenex Hydro RP C18 column.
the retention times for lactic, malic do not change and are rather stable. Succinic acid changes a little (seen as quick as 7.1 min and as long as 9.3 min) and fumaric acid changes drastically from injection to injection (rt 6.7 to 12.5min); sometimes eluting before succinic acid, sometimes after.
mobile phase is 100% 0.1% aq formic acid
samples are in culture broth with the following matrix recipe (1L broth):
2 g of asparagine,
1 g of KH2PO4,
2.5 g of Na2HPO4,
10 mg of MgSO4 · 7H2O,
50 mg of ferric ammonium citrate,
0.5 mg of CaCl2,
0.1 mg of ZnSO4,
0.1 mg of CuSO4, and
0.05% Tween 80
0.5 g of Casitone (Difco) The pH of the medium was 6.8. Carbon source is10mM Glucose.
the rt problem did not exist at first when just using standards and water, but has persisted since first exposure to the above broth. I have tried EDTA sol'n and column wash with 100% acetonitrile and problem continues (even if injecting standards with water).
My questions are this:
#1 what would be a good clean up method to use on the sample prior to injection? ie keep the organic acids, lose the inorganic salts, sugar, and that tween 80 polymer/lipid etc.
#2 does anyone know if the retention time variability is a column contamination issue, or ion-complexing in broth issue? Fumaric acid has a double bond in it. Other acids do not.
#3 are there any ESI compatible methods for organic acid analysis with an Aminex HPX-87H column?
any suggestions or assistance shall be greatly appreciated.
thank you.
I'm experiencing a problem involving Lactic, Malic, fumaric, and Succinic acid using LC/MS ESI with a Phenomenex Hydro RP C18 column.
the retention times for lactic, malic do not change and are rather stable. Succinic acid changes a little (seen as quick as 7.1 min and as long as 9.3 min) and fumaric acid changes drastically from injection to injection (rt 6.7 to 12.5min); sometimes eluting before succinic acid, sometimes after.
mobile phase is 100% 0.1% aq formic acid
samples are in culture broth with the following matrix recipe (1L broth):
2 g of asparagine,
1 g of KH2PO4,
2.5 g of Na2HPO4,
10 mg of MgSO4 · 7H2O,
50 mg of ferric ammonium citrate,
0.5 mg of CaCl2,
0.1 mg of ZnSO4,
0.1 mg of CuSO4, and
0.05% Tween 80
0.5 g of Casitone (Difco) The pH of the medium was 6.8. Carbon source is10mM Glucose.
the rt problem did not exist at first when just using standards and water, but has persisted since first exposure to the above broth. I have tried EDTA sol'n and column wash with 100% acetonitrile and problem continues (even if injecting standards with water).
My questions are this:
#1 what would be a good clean up method to use on the sample prior to injection? ie keep the organic acids, lose the inorganic salts, sugar, and that tween 80 polymer/lipid etc.
#2 does anyone know if the retention time variability is a column contamination issue, or ion-complexing in broth issue? Fumaric acid has a double bond in it. Other acids do not.
#3 are there any ESI compatible methods for organic acid analysis with an Aminex HPX-87H column?
any suggestions or assistance shall be greatly appreciated.
thank you.