randy » Tue Mar 04, 2008 6:44 pm
SammyD, I have the menu in the ADC portion of the method that should allow me to send the TTL signal at a given time before, during, or after a run. I've set the method and the connections up correctly, but I haven't had success in turning off the external detector.
I discovered this morning that the Digital I/O contacts are in the High state by default. That is, with the instrument idling/pumping, the voltage I read from pin 2 and ground is ~4V DC. I set the method to send a TTL Low pulse at 0.1 min after the start of the method. I started the method but didn't see a change, but I found out that a 120msec pulse would not register on the multimeter.
After discovering that the TTL default state is High, I realized that I may not have the detector set to receive the correct signal. The DIONEX has four choices for TTL input: Normal Edge, Normal Pulse, Inverted Edge, and Inverted Pulse. With the Agilent TTL output going from High to Low, I think I'll need the Inverted Pulse setting.
Now I need for the other analyst to finish with the instrument, so that I can test with the new settings.
If they work, I can create a separate method that stops the pumping and sends the signal to turn off the detector. I'll just put this method in the last line of each sequence to turn off the instrument.