Dan » Wed Mar 26, 2008 6:22 pm
I have worked in the cGMP labs at several places and I have noted that these things vary slightly from place to place. It mostly depends on the SOP, but QA and managers can add their own twists to the procedures.
At one place, the manager insisted that the system suitability be run AND checked before any samples are injected. This is not always practical and the SOP didn't require it. So, we usually just proceeded with the entire run and trust that we did everything we could to avoid problems.
I like to start with a test injection (using a standard). I usually have set up the instrument for the entire run; with that test injection first, followed by the system suitability injections and then the rest of the injections. I will check the test injection to see if everything is OK. If not, then I just abort, fix the problem and start over. If time permits, then I check the system suitability results. Otherwise, I will check the results for the run after it is completed, usually the next day as I don't put in 24-hour shifts.
The system suitability is usually based on those 5 (or 6) initial standard injections. With there being acceptance criteria for tailiing, RSD and resolution which are the minimum requirements (see USP and EP). there may be other system suitability criteria such as check standard recovery, detector sensitivity, plates, etc. We did mention some of this before:
http://www.sepsci.com/chromforum/viewtopic.php?t=7976
For issue #2, you can look to the USP where it states that system suitability needs to be checked thoughout the run. Again, I have relied on SOPs to interpert this as I have seen nothing from the USP, FDA, EP or EMEA for clarification. The procedure that I have used (at several different companies) is to check the RSD for the peak areas of the standards throughout the run and that includes the 5 (or 6) system suitability injections.
That's what the SOPs have stated where I have worked. Those SOPs were usually written by the scientists and approved by QA. I am not sure that I would want QA to write those SOPs.
Regards,
Dan