Guard Columns
Posted: Fri Dec 29, 2017 5:00 pm
by madewald
Hello,
I'm try to figure something out. I'm using an Agilent 7890 CG and have recently started to use a guard column. Our software (Atlas) asks for the the column length. Do I need to add in the length of the guard column (5 meters) to the analytical column (30 meters)? This is what I'm currently doing.
Any insight would be appreciated.
Re: Guard Columns
Posted: Fri Dec 29, 2017 6:59 pm
by James_Ball
The instrument calculates column flow from the head pressure and column length. If you add 5m of 0.25 guard column to a 30m 0.25 analytical column you have effectively made a 35m 0.25 diameter column and the instrument software needs to know that to calculate and set the column head pressure to achieve the flow rate you want.
If you add 5m of 0.53 guard column to 30m of 0.25 analytical column, then it would not matter as much since the restriction of the 0.53 guard would only be equal to a few cm of additional column.
Re: Guard Columns
Posted: Fri Dec 29, 2017 7:15 pm
by madewald
Thank you very much.
Re: Guard Columns
Posted: Tue Jan 02, 2018 2:51 pm
by Bigbear
Does the software calculate length if you time an un retained compound?
Re: Guard Columns
Posted: Tue Jan 02, 2018 2:58 pm
by madewald
If you are talking about dead time I don't believe so.
Re: Guard Columns
Posted: Tue Jan 02, 2018 3:23 pm
by James_Ball
I don't recognize the Atlas software, but MSDChemstation has a column calibration section in the software where you can enter the retention time of a non retained peak and it will calculate the column length.