LC-MS Problem, Please help
Posted: Tue Sep 28, 2004 12:48 pm
When I go to 100% ACN (say 5 to 100% ACN in 15 mints and keep 100% ACN for 5 minutes, in a ghradient) in a Trap I always see two predominat ions, 256 and 282 (ESI, +). I normally use SB C8 or C 18 column.
The trap we are using is Agilent SL.
Its is very problematic, specially if the compound of interest elutes out at 100% ACN because these two peaks (282, 256) are so huge that they almost mask other ions.
I have tried cleaning procedure elaborately and it didnt help. The peak is not coming from LC as it is present when I use syringe pump for direct infusion as well.
The agilent engineer is saying it is true universally. Everyone sees those ions when you go go to 100% ACN and keep the gradient for few minutes. Somehow I dont agree with that and I would really really appreciate the readers observation in this regard.
The trap we are using is Agilent SL.
Its is very problematic, specially if the compound of interest elutes out at 100% ACN because these two peaks (282, 256) are so huge that they almost mask other ions.
I have tried cleaning procedure elaborately and it didnt help. The peak is not coming from LC as it is present when I use syringe pump for direct infusion as well.
The agilent engineer is saying it is true universally. Everyone sees those ions when you go go to 100% ACN and keep the gradient for few minutes. Somehow I dont agree with that and I would really really appreciate the readers observation in this regard.