Page 1 of 1

Headspace Restars

Posted: Tue Feb 26, 2008 6:45 pm
by binriaz
Hello,

I am new user of GC and GC is coupled with HEADSPACE and my problem is regarding restarting of HEADSPACE. Whenever headspace is turned on it requires all parameters again and it happened first time in 8 years as my senior informed me. I want to know possible cause and remedies.

Regards,
Kashif

Posted: Tue Feb 26, 2008 8:04 pm
by GasMan
Kashif,

Before anybody can help you, they will need to know the manufacturer and model number of the GC and also the manufacturer and model number of the Head Space Sampler.

Gasman

Posted: Thu Feb 28, 2008 9:41 am
by danicrd
hi...look at backup battery in the headspace and replace it. There is always a battery in the mainboard, and after 8 years i think is normal is gone.
good luck

Posted: Thu Feb 28, 2008 2:38 pm
by binriaz
Thanx a lot for your replies. I will tell you the model as manufacturer is
Perkin Elmer.

Regards,
Kashif Riaz

Posted: Fri Feb 29, 2008 1:10 pm
by binriaz
Model is HS 40XL and GC is also XL autosampler. well my senior confirmed that the problem is due to battery. Thanx a lot