Peak from water at 210 nm
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
			5 posts
							Page 1 of 1
					
	
						I'm using a C8 (4.6 x 250) Inertsil column, the mobile phase is 0.23% perchloric acid/acetonitrile (65:35) premixed, isocratic.  The column temperature is 25C.  injection volume 100 microliters, detection 210 nm. I'm hoping to use this method for determination of related substances however I'm getting a rather broad peak from water (the diluent) at about 10 minutes which will mask some of the related substances.  I've changed HPLC, column, water  and the peak is always there.  Is this typical at 210 nm?  Or is there something strange going on?
					
									- 
								
 - Posts: 1233
 - Joined: Tue Oct 11, 2005 7:20 am
 
						100ul is quite a large injection. and I'm assuming about 1ml/min flow?. 
With a 100ul injection, I'd be worried that some of the sample wasn't soluble in the mobile phase if the sample concentration is high..
What happens if mobile phase is injected, does the peak disappear?.
What happens if 10ul of water is injected, does the peak change?.
What happens if you inject 65/35 H2O: CH3CN without perchloric?>
If it's linked to the injection, I'd look to change that - add acetonitrile and perchloric, or reduce injection volume.
Please keep having fun,
Bruce Hamilton
									With a 100ul injection, I'd be worried that some of the sample wasn't soluble in the mobile phase if the sample concentration is high..
What happens if mobile phase is injected, does the peak disappear?.
What happens if 10ul of water is injected, does the peak change?.
What happens if you inject 65/35 H2O: CH3CN without perchloric?>
If it's linked to the injection, I'd look to change that - add acetonitrile and perchloric, or reduce injection volume.
Please keep having fun,
Bruce Hamilton
- 
								
 - Posts: 15
 - Joined: Wed Jun 08, 2005 1:34 pm
 
						Thank you for your ideas.  Here are the results:
What happens if mobile phase is injected, does the peak disappear?: No, same
What happens if 10ul of water is injected, does the peak change?: Peak is still there only smaller (roughly proportional)
What happens if you inject 65/35 H2O: CH3CN without perchloric?: Peak remains but shape is tailing .
I've calculated that I can afford to decrease the injection volume to about 60 microlitres and still achieve the required sensitivity%. I can't easily change the solvent due to the preceeding procedures. I can't change the wavelength. Any other parameters that you think may be worth looking at please let me know.
Thanks again
									What happens if mobile phase is injected, does the peak disappear?: No, same
What happens if 10ul of water is injected, does the peak change?: Peak is still there only smaller (roughly proportional)
What happens if you inject 65/35 H2O: CH3CN without perchloric?: Peak remains but shape is tailing .
I've calculated that I can afford to decrease the injection volume to about 60 microlitres and still achieve the required sensitivity%. I can't easily change the solvent due to the preceeding procedures. I can't change the wavelength. Any other parameters that you think may be worth looking at please let me know.
Thanks again
- 
								
 - Posts: 2175
 - Joined: Tue Aug 17, 2004 7:59 pm
 
						Try making your own MP & diluent with polished water that has been further filtered through an Empore extraction disk (search my posts featuring "Empore" - I've posted directions etc. at least once). It may help.
					
									Thanks,
DR

				DR

- 
								
 - Posts: 15
 - Joined: Wed Jun 08, 2005 1:34 pm
 
						I didn't have the filters you mentioned but tried purging the water with helium before injection and the peak has disappeared.
Thank you all for your excellent advice.
									Thank you all for your excellent advice.
			5 posts
							Page 1 of 1
					
	Who is online
				In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
 
		
Users browsing this forum: No registered users and 18 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 18 guests
Latest Blog Posts from Separation Science
    Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
    
    
    Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
    
    
- Follow us on Twitter: @Sep_Science
 - Follow us on Linkedin: Separation Science
 