Hi all,
I am running a gradient method on a C18 column 250 x 4.6mm, 5um using a phosphate buffer containg hexane sulfonic acid 1.88g/L pH adjusted to 2.5, with an increasing concentration of acetonitrile with UV detection. As you can see from the embedded picture, (if it works!) I am having some irregular and irreprodicible baseline flactuations which are not acceptable for my application. At the moment I am not considering a detector problem as this is a brand new detector which has been used for only about for 1 month and has always worked perfectly.
I consider myself new to the use of ion paring reagents and since i've never seen such flactuations in normal separations my question is if this could be due to the hexane sulfonic acid since furthermore it is being used in a gradient? Can you please give any suggestions on how can reduce the occurance of such flactuations?
PS: In the above image I consider the bottom chromatogram as representatinve of how it should really look.
Thanks and Regards