1100 Agilent system, baseline drop problem
Posted: Thu Feb 14, 2008 10:31 pm
We have many Agilent 1100 HPLCs in our compnay and we are now recently seeing drastic baseline drops with our validated methods in random injections. Sometimes the base line recovers and sometimes it doesnt. We are seeing this problem using both isocratic and gradient methods.
One of the current changes we did was to switch from Chemstation to Empower 2 to run our HPLCs. I know that Empower and Agilent dont really communicate well, for example there is no function in Empower to see if your degasser is on of off.
I was suspecting outgassing of the mobile phase to cause the baseline drops but if that was true then all of the injections with the drop in baseline should of corrected itself over time, we do not see this in some injections. In some injections the baseline just stays there and doesnt change back to where it should be.
I also suspected that the communication between Agilent and Empower software is not doing its job right, but if that was true then the injections that we saw baseline drops shouldnt of drifted back to the correct baseline.
When we switch over to Alliance systems the system runs perfect.
Any suggestions on how to solve this problem?