Regarding potency of drug
Posted: Thu Feb 14, 2008 6:27 am
by pankajcops
I have one question. We are using Esomeprazole magnesium dihydrate as working standard for analysis of Esomeprazole. The potency of that standard is 94.23 on as is basis . So in calculation we should use Esomeprazole magnesium dihydrate or Esomeprazole magnesium after deducting the water.Can anyone help me in this ?
Posted: Thu Feb 14, 2008 1:10 pm
by JGK
I would use the potency factor of 94.23 (% w/w ?) to calculate the corrected Esomeprazole magnesium concentration from your original weighing.
"As is" implies that the water content has been accounted for in the potency calculation.
If a purity factor was, for example 99.5% w/w (by HPLC/UV/other method) in the documentation, I would adjust as follows:
(Weight -(Weight x % water content)) x purity
Posted: Fri Feb 15, 2008 4:57 am
by pankajcops
If the meaing of as is basis is we have accounted the water content in potency calculation then what we will do in calculation in on dry basis calculation.