Nothing I did seemed to get rid of it, including roaming the lab with a He detector, spraying snoop everywhere, taking all the gas lines apart, checking and rechecking all the vacuum seals, changing regulators, and even eventually dragging the He cyclinder down to one machine and connecting with a single line (we have a large gas line system set up which supplies 2 GCMS's and 2 GC-FID's).
Eventually I tried switching to a different He tank (different Lot#) and the problem vanished, mass 28 and 32 went down to ~2% and 0.5% of He signal during leak check within a few hours. I reconnected the original tank and it came back again.
So my question is this, has anybody else ever encountered a contaminated He tank? I've never ever had a bad tank before and was just curious as to how common this is. Both tanks were from the same supplier but different lot numbers. Next time this will be the first thing I check!