A new colleague asked me if I could run a high/low assay for him. I am unfamiliar with the process, so I asked him for more detail. He said I must first inject samples of differing known concentrations to develop a responce curve. He did not perform the testing so could provide no more information. I currently do not do any quantitive analysis so I am unfamiliar with the process.
I am running Waters 2695's with uv PDA detectors. Can anyone enlighten me or provide a link to learn about the process. I would appreciate any insight you could provide.

Thanks
It sounds like an assay run using a calibration curve ie standards run at 50%-150% target concentration.

You run the sample at two different concentrations, a low concentration for reliable quantitation of the active, and a high concentration for detection and better quantitation of trace levels of degradants or impruties, say at < 0.1%.

The high-low procedure is explained at:

E.L. Inman & H.J. Tenbarge, J. Chromatogr. Sci. 26 (1988) 89.
4 posts Page 1 of 1

Who is online

In total there are 15 users online :: 1 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry