Advertisement

separation in phenyl acetic acid & fluorophenyl acetic a

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Hellow ,,


can any one suggest me in the separation between 2-Fluorophenyl acetic acid and phenyl acetic acid
Hellow ,,


can any one suggest me in the separation between 2-Fluorophenyl acetic acid and phenyl acetic acid
Hi,
For standart solution (K=2-5): Sil C18 (5mkm,C% 14-16), 0.05-0.1M AcOH aq. : MeCN 65- 60:35-40 v/v% or 0.025-0.05М KH2PO4 pH(aq.)=3.5-4.0(H3PO4) : MeCN 60-55:40-45 v/v% UV 254, 230 nm
Thanks for reply,,

I am using 0.05M NaH2PO4.H2O as buffer

Condition

Column : µ-Bondapak
eluent : Methanol : water[0.05M NaH2PO4.H2O , pH 4.0 by 10 % Aq. H3PO4 solution]
Polarity : Methanol [25%] : Buffer [75%]
wavelength : 210nm
Column Temperature : 40°c

Am I in right direction or not please suggest me
Thanks for reply,,

I am using 0.05M NaH2PO4.H2O as buffer

Condition

Column : µ-Bondapak
eluent : Methanol : water[0.05M NaH2PO4.H2O , pH 4.0 by 10 % Aq. H3PO4 solution]
Polarity : Methanol [25%] : Buffer [75%]
wavelength : 210nm
Column Temperature : 40°c

Am I in right direction or not please suggest me
Yes :wink:
But I think that better:
eluent: Methanol : aq. buffer 0.025M pH=3.5 (after mixing the pH is change) pH(eluent)<рКа(analit)!!!
Polarity: Methanol [30%] : Buffer [70%] v/v%
UV : 220nm, 230nm (you must find the absorption spectrum of these molecules)
Col. temp. : ambient or 30
Саn you send me chromatograms to virtulc@gmail.com

I suggest to go to a lower pH. Adjust with phosphoric acid to around pH 2.5 (measured in water). The retention of both analytes will increase a bit. You can compensate for this by adding a bit more methanol. I would also decrease the temperature to 30 degrees.

I also do not see the need for 210 nm. You will get less background (and better signal-to-noise ratio) by going to 254 nM or thereabouts.

Don't worry about any pH changes after you add the organic solvent. This is not important.
5 posts Page 1 of 1

Who is online

In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry