Ethanol for beverages analysis
Posted: Tue Feb 05, 2008 10:48 pm
Hello there, I am new in this forum. So I am learning to use it.
Well I am developing GC analysis for ethanol in a distillery. We are using an Agilent 6850 with FID detector and manual injection. We use an HP-Innowax column 530µm internal diameter.
We want to go as low as possibly detecting the contaminants in ethanol. We are working with: Acetaldehyde, Acetone, Ethyl acetate, methanol, Secbutyl alcohol, propyl alcohol, isobutyl alcohol, 3-pentanol, n-butyl alcohol, n-amyl acetate, 2,4-dimethyl 3 pentanol, isoamyl alcohol, n-amyl alcohol, furfurol, acetic acid, isopropyl alcohol.
We are working with concentrations below 2 ppm. We are using teramyl alcohol as our internal standard (ISTD).
We prepared 4 levels of calibration standards all levels has the same concentration of ISTD. However the chromatograms of the four levels show the ISTD as a growing peak in each level. I mean level 4 is the biggest and level 1 the shorter ISTD. I understand that the ISTD sould show a similar size peak in all levels, but this is not the case.
I have changed liner and septa, have cleaned the FID parts and change de JET, but still thè problem is that the ISTD behaves like stayed above.
I am thinking that the problem is the manifold (the EPC controller) and have requested a change in everything, the EPC module, liner, septa, gold seal and washer, and the collector and jet in the FID.
The same method has been installed in a 6890 in a laboratory next to this one. The only difference is that this is a 6850, and that the 6890 has automatic injection.
Have anyone some idea about that.
Thank you.

Well I am developing GC analysis for ethanol in a distillery. We are using an Agilent 6850 with FID detector and manual injection. We use an HP-Innowax column 530µm internal diameter.
We want to go as low as possibly detecting the contaminants in ethanol. We are working with: Acetaldehyde, Acetone, Ethyl acetate, methanol, Secbutyl alcohol, propyl alcohol, isobutyl alcohol, 3-pentanol, n-butyl alcohol, n-amyl acetate, 2,4-dimethyl 3 pentanol, isoamyl alcohol, n-amyl alcohol, furfurol, acetic acid, isopropyl alcohol.
We are working with concentrations below 2 ppm. We are using teramyl alcohol as our internal standard (ISTD).
We prepared 4 levels of calibration standards all levels has the same concentration of ISTD. However the chromatograms of the four levels show the ISTD as a growing peak in each level. I mean level 4 is the biggest and level 1 the shorter ISTD. I understand that the ISTD sould show a similar size peak in all levels, but this is not the case.
I have changed liner and septa, have cleaned the FID parts and change de JET, but still thè problem is that the ISTD behaves like stayed above.
I am thinking that the problem is the manifold (the EPC controller) and have requested a change in everything, the EPC module, liner, septa, gold seal and washer, and the collector and jet in the FID.
The same method has been installed in a 6890 in a laboratory next to this one. The only difference is that this is a 6850, and that the 6890 has automatic injection.
Have anyone some idea about that.
Thank you.