Separation of diastereomers

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

9 posts Page 1 of 1
Hello
I am working on a method to separate a pair of diasteromers with a view to using the method on a semi-prep scale and am struggling to obtain a separation. The diastereomers are eluted (as a single peak) right at the top of my scouting gradient (100% organic) with both MeOH and MeCN as the organic eluent and on both a C18 and C8 column. I do see some resolution running at Aqueous:Organic 1:1 but it takes about 40 minutes and the peaks are nowhere near well enough resolved. Is there something else I can try to improve things or do I need to use normal phase or a totally different column? Any advice would be very much appreciated.
Thanks.

If you don't have any chiral preparative column, you should start from normal phase column(silica, cyano) and non-polar mobile phase, for diasteromers separation, silica column sometimes gives good resolution.

Diasteromers are usually not possible to separate by changing gradient or buffer (as you may have noticed). It is a matter of finding the right column!

I don't think it is possible to predict the behaviours of different columns, so it is an trial-and-error practise. I have had some luck in the past when using Hypercarb columns for this application, could be worth a try.

Anyone used additives in the mobile phase? I know cyclodextrines are sometimes used in CE.

The Hypercarb column as suggested by Mattias would probably be my first trial-and-error choice. Another possibility is the Cyclobond (cyclodextrin) column. This is normally sold for chiral, but can have good isomer and diastereomer selectivity as well.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Hi,
Trial on chiral columns works out for most of diastereomers,I would suggest to try on chiral AG-P column,it worked out previously for me.
Primesep-D has lots os applications in separating positional isomers(diastereomers).Hope u can try on both columns.
Good luck
Bhaskar Banala

Another possibility is the Luna PFP (PentaFlourylPhenyl) phase which can separate many isomers.
No Tswett

Just to throw in another suggestion: Carbon-clad Zirconia columns. The company is Zirchrom if my memory didn't fail me. Or look up Peter Carr's papers in the literature.

Good luck.

We have been very successful with diastereomers and other closely related products on our ZirChrom-CARB product. I would be happy to supply more information directly. Please feel free to contact me directly with any questions.

Kelly Johnson

support@zirchrom.com
Kelly Johnson
Technical Support & Sales Specialist
ZirChrom Separations, Inc.

There are some companies out there that make C30 phases. I have heard that these can also show good shape selectivity: allowing for separation of things like diastereomers.

Kelly: Have you ever done any comparison of how your ZirChrom-CARB works in comparison to these phases.

Anyone: Have any experience using C30 columns to separate compounds with very similar structures.
9 posts Page 1 of 1

Who is online

In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry