Maximum temperature - Symmetry C18
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
			4 posts
							Page 1 of 1
					
	
						Hi does anyone know where I can find a literature source giving the maximum recommended operating temperature for simple Waters Symmetry C18 columns?
					
									If you are immediately confident in the solution the most likely outcome is the discovery that you have underestimated the problem.
				- 
								
 - Posts: 621
 - Joined: Tue Nov 27, 2007 1:20 pm
 
						Not sure of the actual upper limit.
However, Waters column ovens max out at 60C, so the columns should be OK to that temperature.
									However, Waters column ovens max out at 60C, so the columns should be OK to that temperature.
- 
								
 - Posts: 2916
 - Joined: Mon Aug 30, 2004 10:19 pm
 
						The answer is not simple, since you are dealing with a multi-variable situation. At room temperature, the recommended pH range is between 2 to 8, i.e. you won't see any appreciable changes for about 4 weeks at either pH value. At pH 7 and 50 degrees, you may see appreciable changes already earlier, say at about 2 to 3 weeks. The best stability is obtained around pH 4.5... 10 000 chromatographic runs at 30 degrees without the slightest change in retention.
At what pH are you planning to work, and what will the buffer be?
									At what pH are you planning to work, and what will the buffer be?
- 
								
 - Posts: 7
 - Joined: Fri Sep 15, 2006 8:08 am
 
						Unfortunately the columns are general use columns so the application could vary.  The form that goes with the column asks for a maximum operation temperature to stop users from destroying it or damaging it with a single use.  Based on your comments 50C is a good figure and if they need high temperatures they can ask and we can see whether what they are doing is worth expiring a column so quickly.
					
									If you are immediately confident in the solution the most likely outcome is the discovery that you have underestimated the problem.
				
			4 posts
							Page 1 of 1
					
	Who is online
				In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
 
		
Users browsing this forum: No registered users and 18 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 18 guests
Latest Blog Posts from Separation Science
    Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
    
    
    Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
    
    
- Follow us on Twitter: @Sep_Science
 - Follow us on Linkedin: Separation Science
 