I am running a OI analytical 4552 autosampler with a OI Eclipse 4660, connected to an Agilent 6890/5973 GC/MS. Standard 8260 soil method.
about 2 months ago, I ran a sample with LARGE amounts of 1,1,1-TCE. After running literally about 1000 blanks and changing the entire sample pathway, I had the source cleaned the electron multiplier changed and the column replace as well and the inlet sleeve gold seal, etc... You get the picture. Ever since this work was done, I am having one heck of a time attempting to tune this MS. No matter what I try I cannot get 219 above 131 with my ratios where I want them. Once I get it tuned to the point where I can get BFB to pass it will hold that tune for about 12 hours and then it's a struggle to get a BFB to pass and I'm losing sensitivity as well. For example, Pentafluorobenzene starts with a response of 301951 and 24 hours later I'm getting it responding at 58011.
There is no leak in the autosampler or the concentrator. Ferrules are tight on the column and the gold seal and septa are new and installed properly. I would appreciate any help anybody could offer on this.