What causes MS shutdown after 18min?

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Agilent 6890GC/5973MSD
The MS automatically shut down after 18 min. Could you please help me figure out the possible reasons?
Thanks a lot.

Is the power to the MS getting turned off/is it venting?

Or, is the data not getting collected after 18 minutes?

If it is just not collecting, but the GC continues to run, you probably have a "Detector Off" time of 18 minutes under MS Parameters.

Are there any errors in the log?

wl_yang,

You have a leak and the turbo is not getting up to speed in the specified time.

Best regards.

How often has this happened ? Does it always happen at the same time during a run ?

Peter