is anybody experienced in the verification of the accuracy of the injection volume ? We use an Acquity System (Waters UPLC).
Thank you for every hint
Florian
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
don't think this works for the UPLC cause it's using a loop system, so the volume drawn from the sample is more than actually injectedDear Florian:
You don't have to check the injection volume as part of periodic PQ. This issue was discussed before.
However, you can fill an autosampler vial with water, and weigh it. Then, inject 10 times of the same volume (eg. 50uL), and weigh again after each injection.
W(n) - W(n+1) = Amount of (n+1)th injection.