- Posts: 4
- Joined: Mon Jan 21, 2008 7:34 pm
We have a Varian CP-3800 GC in the lab. The prolbem is the middle pressure couldnot reach the setpoint 60 psi at the 60 mL/min flowrate. Instead, it will stablize at around 40 psi and the system will stay 'not ready'. It's a brand new middle argon type 4 EFC. We just had the varian engineer replaced it a week ago.
I did leak check and found a leak at the middle EFC pencil filter. However, after I fix the leak, the problem still exists. No other leak was detected by using a Restek gas leak detector, model 21-250.
I tried autocalibrate the middle EFC and it made no difference. Although the front type 1 EFC failed to calibrate. But that type 1 EFC has no problem building up pressure.
The argon gas inlet is 80 psi and there is plenty of gas in the tank.
This problem happened before. At that time it was because the backflush solenoid was bad. The varian engieer replaced the backflush solenoid and middle and rear type 4 EFC because they were bad too. It worked for a couple days and failed to build up pressure again. At first the pressure would slowly reach the setpoint if I increased the setpoint 2 psi at a time.
But this trick no longer work now.
This really bothers me. I will really appreciate it if you guys could give me any suggestion on this. Thank you.