Re: C1-C7 aliphatic acids in water with ethylene glycol

Posted: Tue Jan 22, 2008 6:11 pm
by IC_Gal
I am looking for suggestions in analyzing C1-C7 aliphatic acids in a water matrix that has can have % level of ethylene glycol present. We are looking for detection limits in the range of 10 - 50 ppm. What type of SPE and or sample prep would be appropriate? Any column and/or mobile phase suggestions are also welcome.

Posted: Tue Jan 22, 2008 6:21 pm
by Consumer Products Guy
Call Grace/Alltech technical support and ask them. I've been happy with their Prevail Organic Acid column.

Posted: Wed Jan 23, 2008 6:26 am
by XL
Dionex offers the Acclaim Organic Acid column for this application. It is a silica based high-efficiency RP column which is fully compatible with highly aqueous mobile phases. The column housing is non-metallic - PEEK, which would minimizemetal contamination from SST column hardware.

The condition for separating C1 to C7 organic acids:
Column: Acclaim OA, 150x4.0 mm
Mobile Phase: A - MeCN, B - 0.05% phosphoric acid or methanesulfonic acid
Gradient: 0 to 45% A in 11 min and keep for 5 minute (minor modification might be needed)
Flow rate: 1 mL/min
Detection: UV at 210 nm

Here is the typical result you should expect (in Figure 7).
http://www1.dionex.com/en-us/webdocs/25 ... OA_V22.pdf

Due to low wavelength detection, the mobile phase needs to be of carefully prepared with high quality reagents.

If ethylene glycol content is less than 5%, I don't expect too much of trouble. If it does impose some difficulties, using a 250x4.0 mm column should help to some extent.

In case these acids can be analyzed by a RP column due to matrix effect, please let me know so that I can offer an alternative.

Good luck!

Posted: Wed Jan 23, 2008 2:25 pm
by Bryan Evans
We have customers using Unison UK-C18 for organic acids as well.
(using 0.1% H3PO4 as their mobile phase).

Depending on what you are analyzing, you may have analytes eluting with k'<1.
So you should look for dispersion from your LC (e.g. autosampler, tubing, detector).
Also make sure you sampling rate is high enough on your aquisition software.

Posted: Thu Jan 24, 2008 7:32 pm
by IC_Gal
Thanks for all of the suggestions!!