ece » Fri Jan 18, 2008 5:10 pm
Disconnect the column outlet and place in a beaker of solvent and look for bubbles to indicate flow, or measure the flow if you have a set-up. If so, re-install and try a microliter of solvent and see if you get a typical solvent peak, and if successful, go to something simple next. Baby steps. And make sure your detector is operational. Divide and conquer.
That's part of our operational procedure during startup. We do get a signal whether we do a solvent blank or a methane test, but when we do a run with a solvent blank we get very high tailing and noise at the end. I've replaced the jet, I've cleaned all the detector parts, and the inlet parts; so I'm pretty sure it's the column. Could it be the column wasn't conditioned enough?
It sounds like you carrier gas is contaminated. Even if you have a trap to clean ppm amounts of air and water from your carrier gas a leak at the backside of the trap will give you grossly contaminated carrier gas to your GC.
I've leak checked the carrier gas around all the fittings, valves, and traps. The traps also show no sign of contamination and are still active.
leaking septa at the injection port will also allow air into the carrier
The unit will shut down if there's a considerable leak there. Although it takes some degree of pressure drop to activate that feature, I do leak check the inlet whenever I take it apart and put it back together. We replace the septa once a week on top of that.