- Posts: 397
- Joined: Fri Sep 22, 2006 7:36 am
From this morning I'm facing with a problem that I haven't encountered before. For the first time today morning, I used the new air gas cylinder. The previous empty air cylinder was replaced with this new one couple of days before the New Year's Day
When I tried to set the air flow (in the FID parameters) to 450 mL/min, the display of the instrument (PE Clarus 500) showed only cca 380 mL/min air flow. The hydrogen flow was OK (45mL/min).
First thing that I've done was to check for air leaks along the copper tubing from the cylinder to the GC. Everything was OK, no leaking was observed.
Then I started troubleshooting this problem by searching similar problems described in the instrument manuals and this forum also. But I didn't find anything similar to this.
Clarus 500 GC has an electronically pressure control. I wouldn't suspect to the electronics fault because this GC has been recently (a month ago) qualified and verified by PE service.
I did some tests and this is what I got:
1. When I set the FID air flow to 450 mL/min, the actual flow was 377 mL/min
2. When I set the FID air flow to 400 mL/min, the actual flow was 377 mL/min
3. When I set the air flow to 370 and 300 mL/min, the actual flow was 370 and 300 mL respectively.
Then I suspected that maybe some occlusion has occured down the gas lines or (for which I am not convinced) the air cylinder valve cannot support higher flows than 377 mL/min.
The gas supplier is the same. The gas cylinder valve is fully opened and the manometer at the exiting tube shows a pressure above 300kPa.
Everything else seem to work properly, but I'm still confused from this problem. Have anyone ever had similar problems to this.
Your advices are greatly appreciated!
Thanks,
Zoran