Our last column, about $500, was a metal column that we killed in a couple of weeks. We didn’t even get to use it. I’m trying to backtrack the many steps that we did wrong to determine what was likely the root cause of stationary phase breakdown. At the time we didn’t have a leak detector and we didn’t have a replacement ferrule for the one that was cracked, but we installed without leak checking it, b/c we didn’t know. But we did a lot of the following things wrong:
1) If you use a cracked ferrule on the column and run it at high temperatures, can you get air into it and kill the stationary phase?
2) If you have a helium leak somewhere can you get air into the column and kill the stationary phase when you run it at high temperatures?
3) If you have a bent column (like with the metal) that doesn’t go in straight at the inlet, can that affect column performance? If so, how?
4) If you didn’t get a clean cut when you score and break away the column, can that lead to helium leak?
5) If you didn’t get a clean cut when your score and break away the column, and you accidentally get some metal pieces in the column, can that contaminate it?
6) If you install the column backwards, with the guard at the detector end and the activated portion at the inlet side, how does that affect column performance?
7) We were initially told to condition the column at 20 deg. below the maximum temperature the column could handle or 20 deg. below the method’s maximum temperature, whichever was lower, for a period of 20 minutes. This seemed too short to me, until later I found out it is supposed to be 2 hours. I did condition it for 20 minutes the first time. The second time I conditioned it for 2 hrs but it turned out that the column’s stationary phase was pretty much already spent at that point. I cleaned the detector parts in between conditioning runs, but I am concerned that improper conditioning was part of its demise. What is the proper conditioning procedure for a new, 15 m column anyway?
Root cause I would say is #1, not leak checking the system and having a cracked ferrule. What do you think? What are the consequences of all the other ones?